Comparison of a New Test for the Measurement of Resting Whole Saliva with
the Draining and the Swab Techniques
Pia López-Jornet 1
Ambrosio Bermejo-Fenoll 1
J. Vicente Bagan-Sebastian 2
Eliseo Pascual-Gomez 3
1Department of Oral Medicine, University of Murcia, Murcia, Spain
2Department of Oral Medicine, University of Valencia, Valencia,
Spain
3Department of Rheumatology, University of Alicante, Alicante,
Spain
Braz Dent J (1996) 7(2): 81-86 ISSN 0103-6440
| Introduction | Material
and Methods | Results | Discussion
| References |
A quantification method for measuring whole saliva is described. This
whole saliva test (WST) consists of a Whatman paper strip, is easily carried
out, innocuous, low-cost and single use. Due to its characteristics, it
could be considered as the oral equivalent of Schirmer’s tear test. A sample
of 159 healthy subjects (81 males and 78 females; mean age 31.62 years)
participated in this comparative study of this new procedure and two other
tests, the draining and the swab test. Correlation was statistically positive
among the three types of tests.
Key Words: resting whole saliva, salivary flow rate, draining method.
Introduction
The pathological reduction of the salivary flow may cause several problems,
among them an increase of the susceptibility to caries, oral infections
or alterations in mastication, taste, swallowing and diction (Dawes, 1987;
Fox et al., 1987; Mandel, 1989). The sensation of dry mouth, a subjective
symptom, is known as xerostomia (Manthorpe and Axell, 1990). The value
of this symptom, that very frequently does not correlate with the actual
levels of saliva production, is frequently relegated to the subjectivity
of the doctor or the patient. Hence, the importance of the quantification
of saliva production using an objective method. There are several methods
for the measurement of the quantity of saliva that a human being produces
per unit of time, both at resting conditions or stimulating the secretion,
such as for parotid saliva, intra-oral cannulation in the duct, Lashley
cups (and their modifications), and Schneyer’s device. Techniques for the
measurement of whole saliva are usually based on draining to a recipient,
in the re-collection by aspiration or in the increase in weight of an absorbent
material that is chewed or placed somewhere in the mouth (White and Bunting,
1936; Brown, 1970; Stephen and Speirs, 1976; Saunte, 1983; Guter et al.,
1987). However, although used in research, these methods are not normally
used by doctors or dentists in their daily practice. In 1903, Schirmer
described some methods for the functional examination of the lacrimal gland,
based in the wetting of filter paper strips. Schirmer’s Method I measures
the lacrimal secretion at conditions of minimal stimulation. Filter paper
strips are used (0.5 x 3.5 cm) with rounded corners in one end which is
folded and introduced into the conjunctival sac in the intersection of
the medium third and external third of the eye-lid. Five minutes later,
the strips are removed and the length of the moist portion is measured.
Schirmer’s test was conceived to objectively detect the degree of lacrimal
secretion (Prause et al., 1982; Ueda et al., 1983; Yagi, 1986), to evaluate
xerophtalmia, the current diagnostic criteria of Sjögren syndrome
(Manthorpe et al., 1986). The goal of this study is to present an objective
and quantifiable whole saliva measurement test (WST), similar to Schirmer’s
test, based on the wetting of a material, innocuous, easy to reproduce
and carry out, inexpensive and practical.
Material and Methods
The sample was formed by 159 healthy subjects (81 males and 78 females;
mean age 31.62 years). The inclusion criteria to participate in the study
were: (a) not having any systemic or nervous illness or any important physical
defect; (b) not taking any medications or drugs or being under any type
of treatment; (c) not having any symptoms of dry eyes or dry mouth; (d)
not having suffered in the past any type of illness or treatment that could
have an effect on the normal functioning of the salivary gland; (e) pregnant
or nursing women were excluded. Collection was carried out in an air-conditioned
laboratory (mean temperature 20ºC). Tests were carried out between
10:00 and 12:00 a.m. Each subject was instructed to adopt the position
known as “coachman position”: the sitting subject lowers his trunk vertically,
the back is curved and the head hangs forwards slightly. Before starting
the test, the subject swallows the saliva he/she has in the mouth. During
the five minutes of the test, the subject remains relaxed and with eyes
closed. For each subject, three tests were held in the same session, separated
by a resting period of 10 minutes. First, the whole saliva test (WST) which
measures flow rate (cm/min) was carried out. Whatman paper 41 was used
uniformly cut in 1 x 17 cm strips. Then, a millimetred rule was printed,
the first centimeter being left with no print. The strip was introduced
in a polyethylene bag of low density and was sterilized with ethylene oxide.
The non-millimetred centimeter of the strip was extracted from one end
of the bag. Then, this centimeter was folded in a 90º angle and inserted
in the buccal cavity under the tongue, the non-millimetred portion being
in contact with the mouth mucosa. When lips were closed, they lightly contacted
the bag protecting the millimetred portion of the strip from moisture.
The saliva produced accumulated in the lingual vallecula and slowly soaked
the strip. After five minutes, the subject opened his/her mouth, the strip
was extracted and the wet centimeters read (López-Jornet, 1992)
(Figure 1). If during the test, a strip was humified
before the five minutes, the exact time was registered and the rate for
five minutes calculated. We then carried out the draining technique (ml/min)
(Navazesh and Christensen, 1982) and finally the swab weighing test (g/min)(Peck,
1959). We applied Student’s t mean comparison test (P<0.05). We carried
out a simple linear regression analysis and obtained Pearson’s correlation
coefficient; the values were considered statistically significant at P<0.05.
Results
Of the 159 subjects studied, the youngest participant was 5 years old,
the oldest 77, and the mean age was 31.62, with a typical deviation of
13.92. There were 81 males (50.9%) and 78 females (49.2%) in the sample.
With the whole resting saliva test (WST) procedure, the mean value of wetting
of the strip in 5 minutes was 4.328 ± 2.15 cm.The test-retest reliability
(r) was 0.666. The mean values obtained for males was 4.68 cm and for females
3.985 cm. Applying Student’s t mean comparison test of non-paired data
to the results for males and females, we obtained a value of t = 2.02 and
P = 0.0451. With the draining technique, measured in ml, the mean value
for a 5-min period was 1.136 ± 1.153 ml, the lowest value obtained
was 0 and the highest 4.5 ml. With the swab technique, the mean value was
1.156 ± 0.993 g, the lowest value was 0.074 g and the highest 4.989
g (Table 1). A correlation study was carried
out by simple regression of the WST at resting conditions, the draining
test and the swab test, and Pearson’s correlation coefficient was calculated,
as well as the P value. The correlation between the draining and the swab
tests was r = 0.286 and P = 0.0006, The correlation between WST and the
draining test was r = 0.295 and P = 0.0002. The correlation between WST
and the swab test was r = 0.24, P = 0.0031 (Figure
2 A,B,C)
Discussion
As pointed out in the introduction, in spite of the numerous methods for
the collection of saliva, they are rarely used in daily practice. Neither
the general practitioner, nor the specialist, has incorporated into their
usual practice any of the methods used in research for the quantification
of saliva produced per unit of time. Studies that compare methods for quantification
of the salivary flow are scarce. However, our results seem to indicate
that at least the most widely accepted methods are realiable and there
are no important differences among them. The draining test and the swab
test were carried out following the norms given by the classic authors
(Peck 1959; Heintze et al., 1983). Our results with the draining test were
similar to those found by Heintze et al. (1983) but slightly lower than
Navazesh and Christensen (1982). With the swab test, the mean value was
156 g. These values are below those obtained by Peck (1959). Perhaps these
differences could be due to the fact that, in our experiment, subjects
remained very relaxed in the coachman position and with eyes closed, and
therefore, were not submitted to visual stimuli, while the cited authors
allowed subjects to open their eyes. The coachman position was taken from
Schultz’s book “Concentrative self-relaxation” (1969), and we considered
it a good procedure for the normalization of salivary quantification tests.
By simple linear regression, a significant correlation was found between
the draining test and the swab test. This means that when the values of
the draining test increase, the values for the swab test do the same; in
this sense, our results agree with those from other authors (Navazesh and
Christensen, 1982; Heintze, 1983). Similarly, applying a simple regression
analysis, significant relations were obtained between the WST and the draining
test and the WST and the swab test. Thus, when the values of the draining
or the swab test increase or decrease, the values for the WST do the same
(P = 0.0002 for the former and P = 0.0031 for the latter). The WST, as
Schirmer’s test, has proven to be a whole saliva measurement procedure
of easy operation and reproduction, requiring no special equipment. The
strips are inexpensive and do not require much space for storage, as would
be the case for sialometers. The parallelism between Schirmer’s test and
the WST is evident: the same material (Whatman paper 41); the same principle
(the paper strip is soaked by the fluid by absorption); the same time (5
min) and the same measurement carried out in millimetres. These similarities
favor the future use of the WST. Our test also has some small advantages
compared to Schirmer’s test. First, the resting test causes less stimulation,
since the oral mucosa is used to contact with solid elements but the conjunctive
mucosa is not. The stimulus produced when the oral mucosa contacts the
strip and the bag is minimal, since these materials are soft and flavorless.
Other tests have been designed to measure whole saliva flow rate, e.g.,
Saxon test (Kholer and Winter, 1985) which consists in chewing a sterile
sponge, or partial parotid saliva flow such as the test of Davis and Marks
(1986) measuring parotid saliva using a strip of Schirmer’s test The WST,
a new whole saliva quantification test, can be considered as the true oral
equivalent to Schirmer’s test. We could say that WST is to saliva what
Schirmer’s test is to tear. The WST can be used, according to the correlations
obtained, instead of the draining test or the swab test, that require special
equipment and are uncomfortable in their operation, and therefore are not
appropriate for common use in the day-to-day medical practice.
References
Brown CC: The parotid puzzle: A review of the literature on human salivation
and its applications to psychophysiology. Psychophysiology 7: 66-85, 1970
Davis CC, Marks JE: The use of the Schirmer tear test in evaluating
mouth dryness. Dent Hyg 60: 116-129, 1986
Dawes C: Physiology factors affecting salivary flow rate. Oral sugar
clearance and the sensation of dry mouth in man. J Dent Res 66: 648-653,
1987
Fox PC, Busch K, Baum BJ: Subjective reports of xerostomia and objective
measures of salivary gland perfomance. J Am Dent Assoc 115: 581-584, 1987
Guter KD, Beck FM, Blozis GG: A comparison of tecnique for evaluation
of xerostomia. J Dent Res 66: 201, 1987
Heintze U, Birkhed D, Björn HD: Secretion rate and buffer effect
of resting and stimulated whole saliva as a function of age and sex. Swed
Dent J 7: 227-238, 1983
Kholer PF, Winter ME: A quantitative test for xerostomia. The Saxon
test, an oral equivalent of Schirmer test. Arthitis Rheum 10: 1128-1132,
1985
López Jornet P: Estudio clínico de un nuevo test de mensuracion
de saliva global en humanos. Thesis. Universidad de Murcia, Facultad de
Medicina y Odontologia, 1992
Mandel ID: The role of saliva in maintaining oral homeostasis. J Am
Dent Assoc 119: 298-304, 1989
Manthorpe R, Axell T: Xerostomia. Clin Exp Rheumatol 8: 7-12, 1990
Manthorpe R, Oxholm P, Prause JU, Chiodt M: The Copenhagen criteria
for Sjögren’s syndrome. Scand J Rheumatol 61: 19-21, 1986
Navazesh M, Christensen CA: Comparison of whole mouth resting and stimulated
salivary measurement procedures. J Dent Res 61: 1158-1162, 1982
Peck RE: The SHP test - an aid in detection and measurement of depression.
Arch Gen Psychiatry 1: 35-40, 1959
Prause JU, Frost-Larsen K, Isager H, Manthorpe R: Tear absorption into
the filter paper strip used in the Schirmer I test, a methodological study
and critical survey. Acta Ophthalmol 60: 70-78, 1982
Saunte C: Quantification of salivation, nasal secretion and tearing
in man. Cephalgia 3: 159-173, 1983
Schirmer O: Studien zur Physiologie und pathologie der tränenadsorderung
un tränenabfuhr. Albecht von grefes. Arch Klin Exp Ophthalmol 56:
197-291, 1903
Schultz JH: Autorelajación concentrativa. 3rd ed. Científico-Médica
Barcelona, 12-14, 1969
Stephen KW, Speirs CF: Methods for collecting individual components
of mixed saliva: The relevance to clinical pharmacology. Br J Clin Pharmac
3: 315-319, 1976
Ueda T, Nakazawa K, Mori S: Standardization of the Schirmer test in
nomal cases. Nihon Univ J Med 25: 333-342, 1983
White J, Bunting R: A comparison of the chemical composition of stimulated
and resting saliva of caries-free and caries-susceptible children. Am J
Physiol 117: 529-523, 1936
Yagi N: Comparison of thread test of lacrimation to the Schirmer test.
Ann Otol Rhinol Laryngol 122: 3-6, 1986
Correspondence: P. Lopez-Jornet, Hospital Universitario Morales
Meseguer, Clinica Odontologica Universitaria, 2 planta, Cátedra
de Medicina Bucal, C/Marques de los Velez s/n, 30008 Murcia, Spain.
Accepted October 22, 1996
Electronic publication: February, 1997